Fluoroscopy Principles

Back to Topics
Contents
0%

Fluoroscopy Principles

Comprehensive guide to fluoroscopy physics, image intensifiers, C-arm operation, radiation safety, and surgical applications including fracture reduction, hardware placement, and arthrography.

Medium Yield
complete
Updated: 2026-01-16
High Yield Overview

Fluoroscopy Principles

Real-Time Imaging for Orthopaedic Surgery

1/d²Inverse Square Law
7.5-15Pulse Rate (fps)
95%Lead Apron Attenuation
0.5mm PbStandard Lead Equivalent

C-arm Configurations

Standard C-arm
PatternFull-size mobile unit
TreatmentLarge joint, spine, pelvis surgery
Mini C-arm
PatternSmaller, lower power
TreatmentHand, wrist, foot, ankle surgery
O-arm/Cone beam CT
PatternIntraoperative 3D imaging
TreatmentSpine surgery, complex reconstruction
Flat-panel C-arm
PatternDigital detector technology
TreatmentModern standard, better image quality

Critical Must-Knows

  • Inverse Square Law: Doubling distance from source = 1/4 dose. Stand back from the beam.
  • Scatter radiation: Main source of staff exposure. Scatter is highest on the tube (source) side.
  • Pulsed fluoroscopy: Reduces dose vs continuous. Use lowest acceptable pulse rate.
  • Magnification increases dose: Higher mag = smaller field = more dose per area.
  • Collimation reduces dose: Smaller field = less scatter = lower dose to patient and staff.

Examiner's Pearls

  • "
    Scatter is maximum on the tube side - stand on detector (image intensifier) side.
  • "
    Lead apron attenuates approximately 95% of scatter radiation at diagnostic energies.
  • "
    Last-image-hold saves dose by reviewing stored image rather than live fluoro.
  • "
    Mini C-arm for extremities delivers much lower dose than standard C-arm.
  • "
    Pregnant staff should wear wrap-around lead and dosimeter under apron.

Clinical Imaging

Imaging Gallery

Images of patient with right-side L4/5 LDH.
Click to expand
Images of patient with right-side L4/5 LDH.Credit: Open-i / NIH via Open-i (NIH) (Open Access (CC BY))
Fifty-seven year-old female patient who complained of back pain that was unrelieved with medication. A. Thoracolumbar lateral radiograph shows a compression fracture of the T12 vertebra. C-D. Transped
Click to expand
Fifty-seven year-old female patient who complained of back pain that was unrelieved with medication. A. Thoracolumbar lateral radiograph shows a comprCredit: Kim JH et al. via Yonsei Med. J. via Open-i (NIH) (Open Access (CC BY))
Sixty-eight year-old female patient with osteoporosis. A. MRI shows compressive contour changes and marrow edema of the T11 vertebral body. C-D. The transpedicular approach for PVP was performed under
Click to expand
Sixty-eight year-old female patient with osteoporosis. A. MRI shows compressive contour changes and marrow edema of the T11 vertebral body. C-D. The tCredit: Kim JH et al. via Yonsei Med. J. via Open-i (NIH) (Open Access (CC BY))
Oblique needle technique of fluoroscopically guided lumbar fact joint injection.
Click to expand
Oblique needle technique of fluoroscopically guided lumbar fact joint injection.Credit: Schütz U et al. via PLoS ONE via Open-i (NIH) (Open Access (CC BY))

Radiation Safety is the Surgeon's Responsibility

The operating surgeon controls fluoroscopy exposure. You must understand dose reduction principles and apply ALARA. Excessive radiation exposure can lead to deterministic effects (skin burns) and stochastic effects (cancer). Staff dosimetry and radiation safety training are mandatory.

Physics Fundamentals

How Fluoroscopy Works

  1. X-ray Tube: Produces continuous or pulsed X-ray beam
  2. Patient: X-rays pass through patient (differential absorption)
  3. Detector: Image intensifier or flat-panel detector captures transmitted X-rays
  4. Display: Real-time image shown on monitor

Image Intensifier vs Flat-Panel Detector

Detector Technology Comparison

FeatureImage IntensifierFlat-Panel Detector
TechnologyVacuum tube with phosphor screensSolid-state digital array
Image QualityGood, some distortion at edgesExcellent, uniform across field
Dose EfficiencyModerateBetter (higher DQE)
Size/WeightBulky, heavyCompact, lighter
CostLowerHigher (becoming standard)

Radiation Production

Primary Beam

Direct X-ray beam from tube to detector through patient. Highest intensity. Stay out of primary beam path.

Scatter Radiation

X-rays deflected by patient's tissue in all directions. Main source of staff exposure. Maximum intensity on tube side.

Inverse Square Law

Critical Safety Principle

Intensity ∝ 1/distance²

  • Doubling your distance from the source = 1/4 the radiation dose
  • Tripling distance = 1/9 the dose
  • Standing 2 meters away vs 1 meter = 75% dose reduction

Practical application: Stand as far from the beam as possible while maintaining surgical function.

C-arm Operation

Components

C-arm Components

ComponentFunctionLocation
X-ray TubeProduces X-ray beamOne end of C-arm
Detector (II or FPD)Receives transmitted X-raysOpposite end from tube
C-shaped ArmConnects tube and detectorRotates around patient
Monitor/WorkstationDisplays live and stored imagesSeparate from C-arm
Foot PedalControls exposureOperated by surgeon/assistant

Standard Positions

Tube below, detector above patient

  • Standard anteroposterior view
  • Tube typically below table to reduce scatter to staff
  • Most common configuration

C-arm angled to patient axis

  • Obturator/iliac oblique for acetabulum
  • Inlet/outlet for pelvis
  • Careful positioning to avoid beam exposure to staff

Operator Controls

kVp (Kilovoltage)

Controls X-ray penetration and contrast.

  • Higher kVp = more penetrating, lower contrast
  • Auto-adjusted by automatic brightness control (ABC)
  • Manually increased for larger patients

mA (Tube Current)

Controls X-ray quantity (number of photons).

  • Higher mA = brighter image, higher dose
  • Auto-adjusted by ABC
  • Balance image quality vs dose

Pulse Rate

Frames per second for pulsed fluoroscopy.

  • Continuous: 30 fps (highest dose)
  • Standard: 15 fps
  • Low dose: 7.5 fps or lower
  • Lower pulse rate = lower dose

Magnification

Electronic or geometric zoom.

  • Magnification increases dose per unit area
  • Use only when necessary for detail
  • Return to normal mag when possible

Dose Reduction

ALARA Principle

Mnemonic

TDSDose Reduction Strategies

T
Time
Minimize fluoroscopy time. Use pulsed mode. Last-image-hold.
D
Distance
Maximize distance from beam. Inverse square law applies.
S
Shielding
Lead apron, thyroid shield, glasses. Mobile lead shields.

Memory Hook:Time, Distance, Shielding - the three pillars of radiation protection

Practical Dose Reduction

Dose Reduction Techniques

TechniqueHow It HelpsDose Reduction
Pulsed fluoroscopyReduces beam-on time50-75% vs continuous
Last-image-holdReview stored image instead of live fluoroSignificant
CollimationSmaller field = less scatter30-50%
Low magnificationNormal mag uses less doseMag doubles dose
Optimal positioningPatient close to detector, tube awaySignificant
Avoid digital acquisitionStore spots only when needed5-10x per image

Staff Positioning

Where to Stand

Key principle: Scatter is maximum on the TUBE SIDE

  • Stand on the detector (image intensifier) side - less scatter
  • When tube is below table (AP view), scatter goes down and to sides
  • When tube is lateral, stand on the opposite (detector) side
  • Use mobile lead shields when possible
  • Never stand in the primary beam

Radiation Scatter Heatmaps

Radiation dose heatmap for hip replacement and arthroscopy procedures
Click to expand
Radiation dose heatmap showing scatter radiation distribution during hip replacement and hip arthroscopy procedures using a C-arm fluoroscope. The color gradient ranges from blue (lowest radiation) to red (highest radiation). Staff positions are marked with green dots: Surgeon (highest exposure, closest to X-ray source), Surgical assistant, Instrument Nurse (Scrub), Circulating Nurse (Scout), Anaesthetist, and Anaesthetic Nurse. The X-ray source is positioned below the patient table. The right-side legend displays total screening time required to reach 1mSv occupational dose limit, comparing scenarios with and without 0.375mm lead apron protection (e.g., surgeon: 36 hours without lead vs 2,870 hours with lead for hip replacement). This demonstrates the inverse square law relationship and critical importance of lead protection in radiation safety.Credit: Dorman T, Drever B, Plumridge S, et al. Eur J Orthop Surg Traumatol. 2023. PMC10504098. CC-BY 4.0
Radiation dose heatmap for knee surgery in lateral position
Click to expand
Radiation dose heatmap for knee surgery with C-arm fluoroscope in lateral position (X-ray beam horizontal, parallel to floor). The patient is positioned supine with the lateral X-ray beam directed through the knee. Staff positions are marked with green dots. Note the asymmetric scatter pattern with a prominent vertical blue streak indicating radiation shielding provided by the C-arm base and housing. Radiation levels are lower overall compared to hip surgery (more blue/green areas), with the surgeon experiencing 111 cases per year to reach 1mSv without lead protection versus 8,900 cases with 0.375mm lead apron. The lateral beam configuration demonstrates different scatter characteristics than conventional vertical beams used in hip surgery.Credit: Dorman T, Drever B, Plumridge S, et al. Eur J Orthop Surg Traumatol. 2023. PMC10504098. CC-BY 4.0

Mini C-Arm Radiation Safety

Mini C-arms produce dramatically lower scatter radiation compared to standard C-arms:

  • Used for distal extremity procedures (hand, foot, ankle)
  • Operates at lower kVp (typically 52 kVp vs 70-77 kVp for standard C-arm)
  • Lower mA settings (0.06 mA vs 17-20 mA)
  • All staff positions remain in very low radiation zones
  • Most staff require over 10,000 cases per year to reach 1mSv even without lead protection
  • Surgeon position: approximately 5,000 hours of screening time to reach 1mSv
Radiation dose heatmap for foot surgery with mini C-arm
Click to expand
Radiation dose heatmap for foot surgery using a mini C-arm fluoroscope (52 kVp, 0.06 mA). The patient is shown supine with foot elevated. The large black circular object represents the mini C-arm X-ray tube positioned near the foot. Staff positions are marked with yellow dots. The heatmap is predominantly blue, indicating dramatically lower radiation levels compared to standard C-arm procedures. A small yellow glow is confined to the immediate vicinity of the X-ray source. All staff members require greater than 10,000 cases per year to reach 1mSv, even without lead protection, demonstrating the significant radiation safety advantage of mini C-arm units for distal extremity procedures.Credit: Dorman T, Drever B, Plumridge S, et al. Eur J Orthop Surg Traumatol. 2023. PMC10504098. CC-BY 4.0
Radiation dose heatmap for hand surgery with mini C-arm
Click to expand
Radiation dose heatmap for hand surgery using a mini C-arm fluoroscope (52 kVp, 0.06 mA). The patient is positioned supine with hand elevated above the body, visible in the central inset diagram. The black circular mini C-arm X-ray tube is positioned above the hand. Staff positions are marked with cyan dots: Surgeon, Surgical assistant, Instrument Nurse (Scrub), Circulating Nurse (Scout), Anaesthetist, and Anaesthetic Nurse. Very low radiation levels throughout (predominantly blue coloring) with minimal yellow glow around the X-ray source. All staff members require greater than 10,000 cases per year to reach 1mSv without lead protection. This demonstrates minimal scatter radiation during mini C-arm hand procedures, making it the safest fluoroscopic configuration among the five surgical scenarios studied.Credit: Dorman T, Drever B, Plumridge S, et al. Eur J Orthop Surg Traumatol. 2023. PMC10504098. CC-BY 4.0

Radiation Safety

Personal Protective Equipment

Radiation Protection Equipment

EquipmentProtectionRecommendations
Lead apron (0.5mm Pb)Attenuates approximately 95% of scatterMust wear for all fluoro cases
Thyroid shieldProtects radiosensitive thyroidStrongly recommended
Lead glassesProtects lens (cataract risk)Recommended for high-volume users
Wrap-around apronBack protection when turnedRecommended for pregnant staff
Lead glovesHand protection in beamReduce manual dexterity

Dosimetry

Personal Dosimeters

Requirements:

  • Radiation workers must wear dosimeters
  • Badge worn outside apron at collar level (reflects eye/thyroid dose)
  • Second badge under apron optional (reflects body dose)
  • Monthly or quarterly monitoring
  • Dose records maintained

Dose Limits (occupational):

  • Effective dose: 20 mSv/year averaged over 5 years
  • Eye lens: 20 mSv/year (reduced from previous 150 mSv)
  • Extremities: 500 mSv/year

Pregnancy Considerations

Pregnant Radiation Workers

  • Declare pregnancy as soon as known
  • Fetal dose limit: 1 mSv for duration of pregnancy
  • Wear wrap-around lead apron
  • Dosimeter worn under apron at waist level
  • May need role modification if dose limits exceeded
  • Avoid high-dose procedures when possible

Orthopaedic Applications

Fracture Surgery

Fluoroscopic guidance for:

  • Closed reduction assessment
  • Provisional fixation check
  • Final construct verification

Tip: Use last-image-hold to review reduction rather than continuous fluoro

High fluoroscopy use procedure:

  • Entry point identification
  • Guidewire placement
  • Nail passage
  • Distal locking (lateral view critical)

Tip: Perfect circle technique for distal locking reduces fluoro time

Wire and screw trajectory:

  • Cannulated screw over guidewire
  • Down-the-beam view shows screw trajectory
  • Oblique views for intra-articular assessment

Tip: Line up C-arm, insert wire, then check - don't fluoro continuously

Spine Surgery

Spinal Fluoroscopy

Applications:

  • Pedicle screw placement
  • Level localization
  • Disc access for discography/discectomy
  • Vertebroplasty/kyphoplasty
  • Spinal injections

Higher doses: Spine surgery often has highest occupational exposure due to:

  • Multiple levels
  • Lateral views through pelvis
  • Prolonged procedures

Interventional Procedures

  • Joint injections: Confirm needle position before injection
  • Arthrography: Real-time contrast injection visualization
  • Aspirations: Guide needle to fluid collection
  • Nerve blocks: Confirm needle position, visualize contrast spread

Image Quality

Factors Affecting Image Quality

Image Quality Parameters

FactorImproves WithTradeoff
Spatial ResolutionHigh line pairs/mm, small focal spotMore dose, less field
ContrastLower kVp, adequate mALess penetration
BrightnessHigher mA, proper ABC functionHigher dose
Noise (Graininess)Higher mA, lower pulse rateHigher dose

Troubleshooting Poor Images

Image Too Dark

  • Increase mA
  • Check ABC function
  • Reduce patient-detector distance
  • Check for obstruction

Image Too Bright/Washed Out

  • Decrease mA or kVp
  • Remove metallic objects from field
  • Collimate to exclude bright areas

Poor Contrast

  • Decrease kVp
  • Collimate to reduce scatter
  • Position patient closer to detector

Motion Blur

  • Increase pulse rate
  • Stabilize patient/limb
  • Shorter exposure time

Exam Viva Scenarios

Practice these scenarios to excel in your viva examination

VIVA SCENARIOStandard

Radiation Safety Viva

EXAMINER

"During a complex pelvic fracture fixation, the scrub nurse is concerned about radiation exposure. How do you minimize staff dose?"

EXCEPTIONAL ANSWER
I would apply the ALARA principle using time, distance, and shielding. For time: I use pulsed fluoroscopy at the lowest acceptable rate (7.5-15 fps), avoid continuous screening, use last-image-hold to review images, and plan my shots to minimize total fluoro time. For distance: all staff should stand as far from the beam as practical, applying the inverse square law - doubling distance quarters the dose. Staff should stand on the detector side of the C-arm where scatter is lowest. For shielding: everyone must wear lead aprons (0.5mm Pb equivalent attenuates 95% of scatter), thyroid shields, and consider lead glasses. I would position the C-arm with the tube below the patient to direct scatter downward, collimate tightly to reduce the exposed field and scatter, and use mobile lead shields if available. I would also communicate with the team, calling out when screening so staff can step back.
KEY POINTS TO SCORE
ALARA: Time, Distance, Shielding
Inverse square law - doubling distance = 1/4 dose
Stand on detector side (scatter maximum on tube side)
Pulsed fluoro, last-image-hold, collimation
COMMON TRAPS
✗Not mentioning specific techniques
✗Forgetting about collimation
✗Not knowing inverse square law
LIKELY FOLLOW-UPS
"What is your personal dose limit per year?"
"How would you manage a pregnant scrub nurse?"
"What causes the most radiation exposure to the surgeon's hands?"
VIVA SCENARIOStandard

C-arm Positioning

EXAMINER

"You're performing a femoral nailing. Where do you position the C-arm for a lateral view, and where should you stand?"

EXCEPTIONAL ANSWER
For a lateral view of the femur during nailing, I position the C-arm horizontal with the tube on one side and detector on the other. The standard convention is tube posterior, detector anterior - this keeps scatter directed posteriorly away from most of the operating team who are positioned anteriorly. I would stand on the detector (anterior) side of the C-arm, not the tube side, because scatter radiation is maximum on the tube side. If I need to be on the tube side for surgical access, I ensure I'm wearing full PPE and position any mobile lead shield between me and the tube. The image intensifier should be as close to the patient as possible without being in the surgical field - this reduces magnification and improves image quality while reducing scatter. For the lateral view, I need to ensure the C-arm can clear the operating table and any traction equipment.
KEY POINTS TO SCORE
Tube posterior, detector anterior is standard
Stand on detector side (less scatter)
Detector close to patient for best image
Scatter maximum on tube side
COMMON TRAPS
✗Not specifying tube vs detector position
✗Standing on the tube side
✗Forgetting about surgical access requirements
LIKELY FOLLOW-UPS
"How do you get perfect circles for distal locking?"
"What's the difference between image intensifier and flat-panel detector?"
"How does magnification affect dose?"
VIVA SCENARIOStandard

Dose Reduction Techniques

EXAMINER

"A registrar is using excessive fluoroscopy time during a simple distal radius ORIF. What advice would you give?"

EXCEPTIONAL ANSWER
I would advise on several dose reduction strategies. First, use a mini C-arm if available - it delivers significantly lower dose than a standard C-arm for extremity work. Second, avoid continuous fluoroscopy - use single spot images or very short bursts. Third, use last-image-hold extensively - review the stored image rather than screening continuously. Fourth, plan your shots - position the hardware, then take an image, rather than watching the screen while adjusting. Fifth, consider if you need fluoroscopy at all - for many steps like plate positioning, direct vision may be adequate. I would also ensure the pulse rate is set low (7.5 fps is often sufficient for static assessment), avoid magnification unless absolutely needed, and collimate the beam to the area of interest. For training, I would recommend developing mental checkpoints - 'position, then check' - rather than continuous screening. Finally, wearing a dosimeter helps develop awareness of one's radiation use patterns.
KEY POINTS TO SCORE
Use mini C-arm for extremities
Last-image-hold instead of continuous fluoro
Plan shots - position then image
Low pulse rate, avoid magnification
COMMON TRAPS
✗Not mentioning specific practical techniques
✗Forgetting about mini C-arm option
✗Not addressing training/habit formation
LIKELY FOLLOW-UPS
"How much dose reduction does pulsed vs continuous give?"
"What is the acceptable fluoro time for a distal radius ORIF?"
"How do you track your cumulative radiation exposure?"

Fluoroscopy Principles Exam Day Cheat Sheet

High-Yield Exam Summary

Key Physics

  • •Inverse Square Law: Dose ∝ 1/distance²
  • •Scatter maximum on TUBE side
  • •Stand on DETECTOR (II) side
  • •Tube below patient for AP reduces staff scatter

Dose Reduction (TDS)

  • •Time: Pulsed fluoro, last-image-hold
  • •Distance: Stand back, inverse square law
  • •Shielding: Lead apron (95%), thyroid shield
  • •Plus: Collimation, avoid magnification

PPE Requirements

  • •Lead apron 0.5mm Pb equivalent
  • •Thyroid shield recommended
  • •Lead glasses for high-volume users
  • •Dosimeter worn outside apron at collar

Practical Tips

  • •Mini C-arm for extremities (lower dose)
  • •Position then image (not continuous)
  • •Perfect circles for distal locking
  • •Pregnant: wrap-around, waist dosimeter