TRUNNIONOSIS AND TAPER CORROSION
Mechanically-Assisted Crevice Corrosion at Head-Neck Junction
ALVAL SCORE (CAMPBELL)
Critical Must-Knows
- Trunnionosis = mechanically-assisted crevice corrosion (MACC) at head-neck taper
- Risk factors: large heads (36mm+), high offset, long necks, Ti stems, mixed metals
- Cobalt more systemically toxic than chromium (cardiotoxicity, neurotoxicity)
- MARS MRI is investigation of choice - detects soft tissue destruction
- Revision must address soft tissue debridement AND component considerations
Examiner's Pearls
- "Cobalt greater than 7 ppb and Cr greater than 5 ppb are concerning thresholds
- "MoP with large CoCr head on Ti stem = highest risk combination
- "ALTR = Adverse Local Tissue Reaction (umbrella term)
- "ARMD = Adverse Reaction to Metal Debris (same entity, different name)
Critical Exam Concepts
Taper Mechanics Matter
Understand the tribology. Fretting (micromotion) + crevice corrosion (oxygen-depleted environment) = mechanically-assisted crevice corrosion. This is the pathomechanism.
Large Heads = High Risk
Head size is the key modifiable risk factor. Heads 36mm and larger generate greater taper moments and micromotion. Combined with Ti stems, risk is multiplicative.
Metal Ions are Screening
Cobalt and chromium levels guide management. Co greater than 7 ppb is threshold for concern. BUT imaging (MARS MRI) determines tissue damage and surgical decision-making.
Soft Tissue Destruction
ALTR causes extensive tissue necrosis. Pseudotumors, abductor destruction, bone loss. Revision is complex - must debride all necrotic tissue for success.
Taper Corrosion Risk Assessment
| Clinical Scenario | Risk Level | Recommended Action | Key Point |
|---|---|---|---|
| MoP THA with 28mm head on CoCr stem | Low | Standard surveillance | Small head, matched metals = low risk |
| MoP THA with 36mm head on Ti stem | High | Annual metal ions + clinical review | Large head on Ti = high risk combination |
| MoM resurfacing - well-positioned | Moderate | Annual metal ions, clinical review | Bearing surface wear, not taper |
| Symptomatic THA with Co greater than 10 ppb | Critical | Urgent MARS MRI + revision planning | High ions with symptoms = tissue damage likely |
THOLTRisk Factors for Taper Corrosion
Memory Hook:A THOLT of caution with large heads on titanium - the taper takes the toll!
FCGCorrosion Mechanisms
Memory Hook:FCG - Fretting, Crevice, Galvanic - the triple threat of taper destruction!
PUNCHSystemic Effects of Cobalt
Memory Hook:Cobalt can really PUNCH you - systemic toxicity is serious!
Overview and Epidemiology
Definition
Trunnionosis refers to mechanically-assisted crevice corrosion (MACC) occurring at the modular head-neck taper junction of total hip arthroplasty, leading to metal debris release and adverse local tissue reactions (ALTR). It is distinct from bearing surface wear in metal-on-metal articulations.
Epidemiology
- Reported in 2-4% of modular THA overall
- Higher rates with specific designs (up to 10%)
- Male patients may be at higher risk (greater activity)
- Recognition increased dramatically post-2010
- AOANJRR data shows elevated revision rates for certain stem-head combinations
Historical Context
- Modularity introduced 1980s for surgical flexibility
- Initial focus on MoM bearing surface failures
- Taper corrosion recognized as distinct entity ~2010
- MHRA alerts and recalls highlighted problem
- Now affects MoP and CoP combinations too
Pathophysiology
The Corrosion Cascade
Fretting disrupts the protective oxide layer → Crevice environment prevents repassivation → Metal ions released → Local tissue toxicity (ALTR) → Possible systemic toxicity. Understanding this cascade is essential.
Morse Taper Fundamentals
Taper design principles:
- Morse taper creates interference fit via cold welding
- Taper angle varies between manufacturers (5°44' common)
- Taper length affects contact area and stability
- Impaction force critical for initial fixation
Load transmission:
- Axial loads from body weight
- Torsional loads from gait (internal/external rotation)
- Bending moments from offset/neck length
- Larger heads amplify moment arm at taper
Why Large Heads Increase Risk
Moment = Force × Distance. Larger heads increase the lever arm from hip center to taper junction. This amplifies bending moments and torsional stresses at the taper, increasing fretting micromotion. A 36mm head generates approximately 30% more moment than a 28mm head.
Risk Factors
Implant-Related Risk Factors
| Factor | Low Risk | High Risk | Mechanism |
|---|---|---|---|
| Head size | 28-32mm | 36mm+ | Larger head = greater taper moment |
| Stem material | CoCr stem | Titanium stem | Ti has inferior taper performance |
| Metal combination | Matched metals | CoCr head on Ti stem | Galvanic couple accelerates corrosion |
| Neck length | Standard | Long (+5, +10) | Increased lever arm at taper |
| Offset | Standard | High offset | Greater bending moment at taper |
| Taper design | Manufacturer matched | Mismatch or poor design | Taper angle/roughness mismatch |
The worst combination: Large CoCr head (36mm+) on a high-offset titanium stem with extended neck.
High-Risk Combinations to Remember
Titanium stem + Large CoCr head + High offset = Maximum risk. Australian registry data confirms these combinations have the highest revision rates for ALTR. Avoid this combination or implement surveillance protocols.
Clinical Presentation
Local Symptoms
- Groin pain (may be insidious onset)
- Hip pain with activity
- Clicking, squeaking, or grinding
- Swelling (pseudotumor)
- Progressive weakness (abductor destruction)
- Sense of instability
- Recurrent dislocation
Systemic Symptoms
- Fatigue and malaise
- Cognitive changes (memory, concentration)
- Peripheral neuropathy
- Visual or hearing changes
- Palpitations or dyspnea (cardiomyopathy)
- Hypothyroidism symptoms
- Depression
Clinical Stages
Progression of Trunnionosis
Metal ion release begins. Patient asymptomatic. May only be detected on surveillance bloods. Tissue changes minimal.
Groin or hip pain develops. May be mistaken for other causes. Early soft tissue changes on MRI. Metal ions typically elevated.
Pseudotumor formation. Abductor destruction. Bone loss possible. Significant functional impairment. High metal ions.
Cobaltism develops. Cardiac, neurological, thyroid manifestations. May be irreversible. Urgent revision required.
Red Flags for Cobaltism
Systemic cobalt toxicity requires urgent action. Signs include: unexplained cardiomyopathy, peripheral neuropathy, visual/hearing loss, cognitive decline, hypothyroidism. Cobalt levels often greater than 20 ppb. Can be fatal if not addressed. Echo and cardiology referral essential.
Investigations
Serum Metal Ion Testing
| Ion | Normal | Concerning | Action Threshold |
|---|---|---|---|
| Cobalt (Co) | Less than 1 ppb | Greater than 4 ppb | Greater than 7 ppb = imaging required |
| Chromium (Cr) | Less than 1 ppb | Greater than 4 ppb | Greater than 5 ppb = imaging required |
| Co:Cr ratio | Approximately 1:1 | Greater than 2:1 | High ratio suggests taper source |
Interpretation pearls:
- Co:Cr ratio greater than 2:1 suggests taper corrosion (rather than bearing wear)
- Levels can fluctuate - trend is important
- Some patients develop ALTR with low ion levels (hypersensitivity)
- Whole blood samples required (not serum alone)
MHRA guidance (UK) - widely adopted:
- Co or Cr greater than 7 ppb = concern, requires imaging
- Annual surveillance for at-risk implants
Investigation Protocol for Symptomatic THA
1. Serum metal ions (Co and Cr) → 2. Plain radiographs → 3. MARS MRI if ions elevated or symptoms persist → 4. Aspiration to rule out infection before surgery. Do not skip the aspiration - occult infection can coexist.
Differential Diagnosis
Painful THA - Differential Diagnosis
| Diagnosis | Key Features | Distinguishing Investigation |
|---|---|---|
| Trunnionosis/ALTR | Elevated metal ions, pseudotumor on MRI | Metal ions + MARS MRI |
| Periprosthetic infection | Fever, elevated CRP/ESR, positive cultures | Aspiration, WBC count, culture |
| Aseptic loosening | Start-up pain, progressive, radiolucent lines | Serial X-rays, bone scan |
| Instability | Recurrent dislocation, sense of giving way | Clinical exam, X-ray, CT for version |
| Periprosthetic fracture | Acute pain after trauma, may be subtle | X-ray, CT if needed |
| Referred pain (spine) | Back pain, radicular symptoms | Spine imaging, selective injection |
| Iliopsoas impingement | Anterior groin pain, worse with flexion | CT (cup position), lidocaine injection |
ALTR vs Infection
Must distinguish ALTR from infection before revision. Both can present with pain and elevated inflammatory markers. Aspiration is MANDATORY. ALTR fluid is typically dark/metallic but sterile. Infection requires completely different management.
Management
Non-Operative Surveillance
Indications for surveillance (no immediate surgery):
- Asymptomatic patient with at-risk implant
- Low-grade metal ion elevation (Co less than 7 ppb)
- No significant ALTR on MARS MRI
- Well-fixed, well-positioned components
Surveillance protocol:
- Annual clinical review
- Annual metal ions (Co and Cr)
- MARS MRI every 1-2 years if ions elevated
- Patient education on warning symptoms
Upgrade to revision if:
- Symptomatic progression
- Rising metal ion trend
- ALTR progression on MRI
- Systemic symptoms develop
Soft Tissue Debridement is Critical
The single most important factor for revision success is complete debridement of necrotic tissue. Retained necrotic/reactive tissue leads to persistent inflammation, poor healing, and high complication rates. Be aggressive with debridement.
Surgical Technique Considerations
Surgical Approach
Extensile approach required:
- Posterior approach with extended capsulotomy
- May need trochanteric osteotomy for access
- Must visualize extent of pseudotumor
- Direct lateral may limit access posteriorly
Key exposure steps:
- Identify and protect sciatic nerve (often displaced by pseudotumor)
- Evacuate pseudotumor contents carefully
- Complete capsulectomy and synovectomy
- Debride all necrotic tissue back to bleeding margins
Titanium Adapter Sleeve
Ti sleeve/adapter devices allow retention of a well-fixed stem with damaged trunnion. The sleeve covers the corroded taper and provides a new surface for head impaction. This avoids stem revision morbidity in selected cases. Check manufacturer compatibility.
Complications
Complications of ALTR and Revision Surgery
| Complication | Rate | Risk Factors | Management |
|---|---|---|---|
| Dislocation | 10-25% | Abductor loss, revision surgery | Constrained liner, bracing, re-revision |
| Infection | 5-10% | Necrotic tissue, prolonged surgery | Debridement, antibiotics, staged revision |
| Nerve injury | 2-5% | Scarring, pseudotumor displacement | Careful dissection, may recover |
| Persistent symptoms | 15-30% | Incomplete debridement, tissue damage | May need re-revision |
| Abductor insufficiency | 20-40% | Pre-existing damage from ALTR | Gait aids, reconstruction options limited |
| Fracture | 2-5% | Bone loss, osteolysis | May need additional fixation |
High Complication Rate
Revision for ALTR has higher complication rates than routine revision THA. Dislocation rates of 10-25% reflect abductor deficiency. Counsel patients appropriately. May require constrained components, brace, or assistive devices long-term.
Outcomes and Prognosis
Prognostic Factors
Good Prognosis:
- Early diagnosis before extensive tissue damage
- Intact or recoverable abductors
- Good bone stock
- Low metal ion levels
- Complete surgical debridement achieved
Poor Prognosis:
- Delayed diagnosis with extensive destruction
- Abductor loss/atrophy
- Significant bone loss
- Very high metal ion levels (cobaltism)
- Systemic toxicity (cardiac, neurological)
- Multiple prior revisions
Outcomes Data
Re-revision rates after ALTR revision are approximately 15-25% at 5 years - significantly worse than revision for other indications. Dislocation is the most common reason for re-revision. Early diagnosis and meticulous surgery optimize outcomes.
Evidence Base and Key Studies
Langton DJ - Taper Corrosion in Large-Head MoM THA
- First systematic description of taper corrosion as distinct from bearing wear
- Larger head sizes associated with increased corrosion
- High offset stems increased risk
- Titanium stems had worse taper performance
AOANJRR - Metal-on-Metal Hip Arthroplasty Analysis
- MoM bearing has highest revision rate of all bearing couples
- Large-head MoM THA has higher revision than resurfacing
- ALTR accounts for majority of MoM revisions
- Certain stem-head combinations have significantly elevated risk
Hart AJ - MARS MRI for ALTR Detection
- MARS MRI highly sensitive for pseudotumor detection (80%+)
- Solid lesions associated with worse tissue destruction
- MRI findings correlate with intraoperative damage
- Metal ions alone not sufficient for diagnosis
Mistry A - Outcomes of Revision for ALTR
- Re-revision rate 16% at 5 years
- Dislocation most common complication (18%)
- Worse outcomes with delayed diagnosis
- Complete debridement associated with better outcomes
Exam Viva Scenarios
Practice these scenarios to excel in your viva examination
Scenario 1: Surveillance of At-Risk Implant
"A 58-year-old man is 5 years post primary THA with a 36mm cobalt-chrome head on a titanium uncemented stem. He is currently asymptomatic but was referred for surveillance due to implant concerns. How would you assess and counsel this patient?"
Scenario 2: Symptomatic ALTR Requiring Revision
"A 62-year-old woman presents with progressive right hip pain and weakness 7 years after THA. She has a 38mm metal-on-polyethylene bearing on a titanium stem. Her cobalt is 18 ppb and chromium is 6 ppb. MARS MRI shows a large posterolateral pseudotumor with abductor muscle atrophy. How would you manage this patient?"
Scenario 3: Systemic Cobalt Toxicity
"A 55-year-old man presents with fatigue, hearing loss, and peripheral neuropathy. He had bilateral THA 6 years ago (large head MoM). His cobalt level is 85 ppb. MARS MRI shows bilateral pseudotumors. His cardiologist has diagnosed new cardiomyopathy. How would you approach this complex case?"
Australian Context
AOANJRR Data
- Australia has comprehensive registry data on THA outcomes
- MoM THA revision rates significantly higher than other bearings
- Registry identified high-risk implant combinations early
- Data supports current surveillance recommendations
- Specific stem-head combinations flagged for monitoring
TGA Guidance
- TGA has issued alerts on MoM hip replacements
- Surveillance recommendations align with international guidance
- Patient notification requirements for at-risk devices
- Reporting of adverse events to TGA encouraged
- Database of recalled/monitored devices
AOANJRR Findings on Metal-on-Metal THA
The Australian Orthopaedic Association National Joint Replacement Registry has provided world-leading data on hip arthroplasty outcomes. Key findings relevant to trunnionosis include:
Bearing surface data:
- MoM bearings have consistently shown the highest revision rates of all bearing couples
- Large-head MoM THA (36mm+) performs worse than resurfacing
- Ceramic-on-ceramic and ceramic-on-polyethylene have lowest revision rates
Stem-head combinations:
- Specific combinations of titanium stems with large CoCr heads flagged
- Revision for ALTR/metal-related pathology increasing as a percentage
- Registry data has informed surgeon implant choice and surveillance protocols
Clinical implications for Australian practice:
- Surgeons should check registry data before implant selection
- Patients with at-risk implants should be in surveillance programs
- Registry reporting assists ongoing monitoring of implant performance
Patient Notification
Patients with TGA-flagged implants should be notified and enrolled in surveillance programs. This includes patients with large-head MoM THA and certain high-risk conventional THA combinations. Maintain accurate implant records for all patients.
TRUNNIONOSIS AND TAPER CORROSION
High-Yield Exam Summary
Definition
- •Mechanically-assisted crevice corrosion (MACC) at head-neck taper
- •ALTR = Adverse Local Tissue Reaction (umbrella term)
- •ARMD = Adverse Reaction to Metal Debris (same entity)
- •ALVAL = histological pattern of metal hypersensitivity
Risk Factors (THOLT)
- •Titanium stem (Ti has worse taper performance)
- •Head size large (36mm+ increases taper moment)
- •Offset increased / Long neck
- •Lateral offset designs
- •Taper mismatch / Mixed metals (CoCr on Ti)
Corrosion Mechanisms
- •Fretting: micromotion disrupts oxide layer
- •Crevice: oxygen-depleted environment accelerates corrosion
- •Galvanic: dissimilar metals create electrochemical gradient
- •All three combine in MACC
Metal Ion Thresholds
- •Cobalt greater than 7 ppb = concern, imaging required
- •Chromium greater than 5 ppb = concern
- •Co:Cr ratio greater than 2:1 suggests taper source
- •Very high levels (greater than 20 ppb) = check for systemic toxicity
MARS MRI Findings
- •Pseudotumor (cystic or solid)
- •Muscle atrophy (abductors)
- •Fluid collections
- •Bone involvement/osteolysis
Revision Principles
- •Complete debridement of necrotic tissue (most important)
- •Ceramic head to eliminate metal ions
- •Smaller head size (32mm)
- •Ti adapter sleeve if stem well-fixed but taper damaged
- •Consider constrained liner if abductors deficient
Systemic Cobalt Toxicity (PUNCH)
- •Peripheral neuropathy
- •Unexplained fatigue
- •Neuro-ocular toxicity (vision/hearing)
- •Cardiomyopathy (can be fatal)
- •Hypothyroidism